Hello-
The reason you are seeing the ghosting of the counter output signal over your temperature channel is likely related to the method you are using to cable to your SCXI module. The SCXI-1125 is capable of operating in parallel mode, so this means that all channels are physically connected to the DAQ board channels in the module (i.e. ai1 on the module connects to ai1 on the DAQ board).
When running the module in multiplexed mode all SCXI readings are multiplexed to ai0 on the MIO board, but the physical connection between ai1 on the SCXI module still exists. This creates a situation where the two signals are double-driving the input line. To sum up the workaround, you basically either need to use an SCXI-1346 for chassis communication (instead of an SCXI-1349 that you are probably using now) or cable to a non-parallel mode capable module.
The fact that you are not seeing the problem in VB is probably fitting with your assumption that the signals are not being properly output. If you connect your clock signal to a different ai channel that is not being driven by a TC then the problem should be avoided.
Hopefully this helps-
09-19-2005 12:39 PM - edited 09-19-2005 12:39 PM
Message Edited by SCXI and MS 2k3-VB.NET on 09-19-2005 12:40 PM
09-19-2005 12:42 PM
09-21-2005 05:22 PM
I have a problem again. If I run the VB application and then stop it, turn the SCXI chassis off, turn it back on and then run the program again the temperatures are almost twice what they should be. If I then stop VB, open LabVIEW and run it the temperatures go back to normal. Then I can run VB again and see the correct temperatures.
Is there some way I need to reset something?
09-22-2005 11:33 AM
Hello-
The SCXI system does not automatically recover from power downs during operation. You will need to call the DAQmxResetDevice to reset each SCXI module before you should expect to receive "good" readings again after you power down.
09-22-2005 12:52 PM
Can you give me the exact line of code? I searched for that functionin the measurement studio help files and didn't find it. The only thing I found was a device reset.
Thanks.