Sampling rate issues in Labview
Hi Everybody,
I'm trying to measure an oscillating pressure signal using an Endevco 8530C through a NI 9237 module.
For start, I'm trying acquire 1024 samples in a rate of 1024/sec, which should take a full second to complete. For some reason when I use the DAQ assistant, it only measures for about 0.6 second instead of a full second. Changing the number of samples or the rate does not help, and it still doesn't take the full measurement.
Nevertheless, when I do the same thing using the DAQmax, it works fine.
Any ideas as to what I'm missing here?
Two screenshots are attacked.
Thanks!
(3,215 Views)
Accepted by topic author THERMAN
Your problem is you need to read the data sheet.  On page 8 you will see a section on Data Rates.  In it, you will see that the slowest the 9237 can sample at is 1613 S/sec.  So that is the rate the task is actually running at.  So (1024 S)/(1613 S/sec) = 0.635 sec, which corresponds with what you are seeing with the DAQ Assistant.  Your other graph is showing sample numbers, not time, so that is not a valid comparison.
(3,193 Views)