10-30-2019 05:17 AM
Dear all,
my question is the following:
I want to control a HM8118 with Labview. Driver is installed and working.
BUT: If i run the included example VI "HAMEG HM8118 Acquire Example.vi" (which i attached as picture here), the output data is always 0. If i run the sub-VI "read.vi", the measurement output shows the correct value. What's wrong? In my self written VI, the output is always NaN.
Best,
Alex
10-30-2019 06:04 AM
Hi Alex,
did you set your computer to English regional settings, i.e. using the point as decimal seperator?
This is an usual problem with default device drivers, especially in Germany…