Problems with Reading Signal From a Omega PX409-050G5V

Hi everyone,

I am using an Omega PX409-050G5V pressure transducer (0-50 psi) to measure the pressure of a gas tank, as gas flowing into the tank.

I power the pressure transducer with a Model T-2 Blind Power supply of 15 V (Sierra Instrument Inc.).

The transducer is connected to an NI 9219 module. 

I have trouble reading the voltage signal from LabView. When I flow a gas into the tank, I don't see any increase in the pressure (or voltage). I don't know what I have done wrong.

The transducer is located at the inlet of the tank. The initial pressure of the tank is atmospheric pressure. The gas flows from a gas cylinder with pressure of 40 psig.

Could you please help me with this problem?

Pressure_Reading.png

0 積分
1 條訊息(共 14 條)
5,325 檢視
First, let's get LabVIEW out of the way. Can you read signal changes in MAX?

Mike...

Certified Professional Instructor
Certified LabVIEW Architect
LabVIEW Champion

"... after all, He's not a tame lion..."

For help with grief and grieving.
0 積分
2 條訊息(共 14 條)
5,308 檢視

No, I don't  see any changes in MAX as I flow gas into the tank. 

Does it mean that I have a bad connection or the pressure transducer broke?

0 積分
3 條訊息(共 14 條)
5,291 檢視
Or you are looking at the wrong channel
Or it is hooked up wrong
Or the card is not working (have you done a self test?)

Mike...

Certified Professional Instructor
Certified LabVIEW Architect
LabVIEW Champion

"... after all, He's not a tame lion..."

For help with grief and grieving.
0 積分
4 條訊息(共 14 條)
5,287 檢視

I believe that I had the right channel.

I did the self test, and it was successful.

I have checked all the wires again and again, and confirmed it with one of my colleagues in my lab. 

That is why I am lost as why it is not working.

 

Khanh

0 積分
5 條訊息(共 14 條)
5,279 檢視

Hi

 

Have you tried to create a task and verify the Connection Diagram tab (MAX) just to double check if everything is fine? Have you tried to use another computer?

 

Regards

0 積分
6 條訊息(共 14 條)
5,084 檢視

Yes, I did. As I hook the transducer to a gas tank, the output voltage remains zero despite the tank pressure is increasing. My guess is the transducer itself is damaged.

0 積分
7 條訊息(共 14 條)
5,057 檢視

I see. Yes, could be =( 

0 積分
8 條訊息(共 14 條)
5,052 檢視

Exactly how do you have everything connected? Where is the common/return/ground of the power supply connected? Which pins on the 9219 are connected to what? where are the wires on the PX409 connected?

 

Lynn

0 積分
9 條訊息(共 14 條)
5,036 檢視

The Omega 5V transducers are 3-wire devices. You supply power (12-30 V, if memory serves) on one line, there's a return line which should be connected to ground on the power supply _and_ to a ground connector on the data acquisition board, and finally a signal line which gets connected to the analog input connection.

 

Does this match how your wires are arranged? 

0 積分