Oxford Mercury series instruments : Works with NI-MAX but fails with Labview
  I am trying to interface an Oxford Instruments "Mercury Series" power supply for superconducting magnets, I am using Labview 2013 and an NI USB-GPIB-HS.
The problem I have is the following:
1.  After connecting the instruments I run NI-MAX (ver 14.0) to check that the instruments are recognized by the GPIB-bus. Everything goes fine, the instruments are detected and I can communicate with it by typing commands in the 488.2 communicator. Read and write both work fine.
2. Then I try to write my own code in Labview to run the instrument.  The standard 488.2 commands Send/Recieve or GPIB-Write anf GPIB-Read do not work.
There are other instruments on the same bus also - which work fine with both NI-MAX as well as my vi-s from Labview.  I have tried doing the "obvious" things
like intialising the GPIB before sending any command and then do a Device Clear for the specific instruments before sending other commands.
3. So the question is : What can be the reason for an instrument to respond correctly via NI-MAX but fail to do when addressed from Labview with essentially
similar basic commands?
I have tried tracing the origin of the errors via I/O trace (NI-spy), but they didn't help in locating the precise problem.
Would be grateful if somebody could suggest an answer.
Have other people faced this before?
Thanks-
Kanti
Message 1 of 8
(4,114 Views)
You have not provided your code, exactly what you sent in MAX, or the I/O Trace file. I'll make a guess that it's an older instrument that requires a cr or lf termination.
Message 2 of 8
(4,096 Views)
The instruments are actually newer  (rather than an older version). You are correct that they need an extra "\n" after the command string - which I had provided.
The command tryped in NIMAX was simply  "*IDN?\n"   in response to which  the instrument sends its identification string correctly.
All other commands also worked properly.
I then tried the same in Labview using   "Send" and "Recieve". The Recieve generates the error.
I am attaching the trace file (converted to txt) and the VI. Line 10-14 has the error.
Message 3 of 8
(4,084 Views)
You need to change the data string to '\' Codes Display before typing \n will generate a line feed.
Lynn
Message 4 of 8
(4,075 Views)
And in addition, any GPIB instrument that requires a termination character besides EOI, is not compliant with IEEE-488.2 so no matter how 'new' you think the instrument is, the firmware is incredibly old.
Message 5 of 8
(4,056 Views)
Checking the "\ Codes Display" fixed the problem. Labview now communicates correctly with it.
Thanks to Dennis_Knutson & Lynn!
-Kanti
Message 6 of 8
(4,014 Views)
Why does the idn query return Stanford Research instead of Oxford Instruments?
Message 7 of 8
(4,002 Views)
I had a "Stanford Research" instrument also on the same bus - I was also sending a query to it as a check that the bus itself was not hanging.
So the IO-trace captured that also.
Message 8 of 8
(3,984 Views)
