Number of Data Samples

Solved!
Go to solution

I have set my scope to 40Ms/s and a time span of 50uS resulting in a expected 2kpts waveform on the screen. 

When I then get the waveform with labview I got 32kpts. I don't understand why. 
Thanks for any help/tips you can share with me.

 labview-blockdiagram.PNG

labview-frontpannel.PNG

Screenshot-wss.jpg

0 Kudos
Message 1 of 3
(2,683 Views)

I would imagine that it gets set in the Read Waveform vi that was not attached.  32K just so happens to be the max value of an I16 (+32767).

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 2 of 3
(2,651 Views)
Solution
Accepted by topic author hansvz

The vi was attached to the scope. 
I got the answer from someone in the labview community. The scope interpolate the extra point by default. When I switched interpolation off I got the expected 2k points.
(:ACQuire:INTerpolate off)


0 Kudos