NI 9205 voltage dropping to 0 during measurement
Hello everyone,
I'm a self taught LabView user so I'm not too sure that I'm doing this correctly.  But I am using a NI 9205 with my cDAQ system to measure a voltage coming from a mass flow controller.  I have the V+ wired into ai5 and the ground into ai13.  During my measurements, the voltage will constantly change between the expected V_out and 0 V.  I connected the mass flow controller to a DMM and see a constant voltage so I don't believe that it is due to the mass flow controller.
I have attached my code that I'm using to measure and a picture of what I see. Anyone see this issue before?
Message 1 of 8
(4,358 Views)
Hi Breathometer,
what about setting a sample rate for your DAQmx task?
What about reading blocks of samples using a "n samples" mode of DAQmxRead?
What about using a "differential" channel mode when wiring ai5 and ai13?
Why do you use a "n channels" mode, when your task is initialized to use only ai5?
Do you know this one?
Message 2 of 8
(4,339 Views)
Hi GerdW,
Thanks for the reply.  I'll give your suggestions a try! As for the N samples, I saved it on the wrong setting.  I first tried 1 channel 1 sample but I saw the same thing so I tried switching it and forgot to change it back.  I have also tried using a differential sampling mode but that didn't seem to fix my issue either.
Best,
Ellery
Message 3 of 8
(4,330 Views)
Hi GerdW,
I added a DAQtiming VI before my DAQread and set it to 10Hz.  Now it just has ~10sec of expected voltage readings and then it drops back down to 0V for another ~10 sec and so on and so on.  If I change it to 5Hz, then the cycle is about 20 sec expected, 20 sec @ 0V
Best,
Ellery
Message 4 of 8
(4,313 Views)
Hi Ellery,
when changing your VI you should attach it again…
Message 5 of 8
(4,300 Views)
Oh sorry about that will attach.
Message 6 of 8
(4,297 Views)
A few thoughts:
  • If you are wired for differential, the input terminal configuration needs to be set to differential
  • The minimum and maximum should be set as well
  • The DAQmx timing setup should be outside the while loop, there is no reason to set it every iteration of the loop
  • Your chart has the x axis labeled as time, but it's really # of data points collected (which if you know your rate can be easily scaled to time, but I've seen it throw a few people off)
Message 7 of 8
(4,282 Views)
Accepted by topic author ProteinSimple
Hi all,
I appreciate all your input in trying to help me figure out the issue.  Looking online more, I found out that it was a wiring issue.  Because the module is isolated, the AI channels are floating when compared to chassis ground.  So besides having the AI+ and AI- wired, I needed to wire the COM channel to chassis ground as well.
Hope this helps others with similar issues.
Best,
Ellery
Message 8 of 8
(4,224 Views)
