Measuring current on a Voltage Rail > 10V

Solved!
Go to solution

 

I am attemting to measure the current and voltage on a 15V power rail.  (Here is a copy of my circuit attatched) I am using a USB-6225 DAQ and Lab View,

 

 I have my device under test connected to a power supply. I have connected two differential channels to the DAQ, one for the current and one for the voltage.

 

I understand that the DAQ can not handle more than 10V so I use a voltage divider to half the voltage. When I run the measurements, the current reported is incorrect. It shows 10A when it should be about 500mA.

 

When I reduce the voltage on the power supply to 10V, it all works fine, so I am guessing its got something to do with the 10V limit.

 

Please can you advise how I can connect this up correctly?

 

0 Kudos
Message 1 of 6
(2,737 Views)

Why are you posting a hardware issue to a software forum?

Bill
CLD
(Mid-Level minion.)
My support system ensures that I don't look totally incompetent.
Proud to say that I've progressed beyond knowing just enough to be dangerous. I now know enough to know that I have no clue about anything at all.
Humble author of the CLAD Nugget.
0 Kudos
Message 2 of 6
(2,723 Views)

R(Sense) should be within 10 volts of the DAQ chassis.  Thats all the common mode rejection you are likely to have on a DAQ device.  (But check your device specification sheet ---- don't trust me RTFM!)

 

if you do not have a common "Ground" between the DAQ device and the supply you are monitoring  you will need an opto-coupler (But any first year EE student could tell you that)

 

As an Asside.:  Who the hell ever started the convention of stuffing R (sense) on the + side of a circuit anyhow?  Use a regulated supply with "Remote sensing"  if you really care about the DUT power consumption.


"Should be" isn't "Is" -Jay
0 Kudos
Message 3 of 6
(2,711 Views)
Solution
Accepted by topic author rodwatt

Do it this way.

DAQ Set Up.jpg

========================
=== Engineer Ambiguously ===
========================
Message 4 of 6
(2,697 Views)

Great, that worked!

 

Thank you very much for your constructive feedback and suggestion, much apprecited!

0 Kudos
Message 5 of 6
(2,678 Views)

High side current sensing does prevent ground disturbances for sensitive or high speed PWM/Inverter circuits.

I think it can be used for additional diagnostic info that low side can't provide, can't recall.

 

EDIT: Forgot- with high side you can use multiple sensors to monitor current of multiple loads on a common ground buss.

 

Never had a need to use it though.

 

-AK2DM

~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~
0 Kudos