I'm also working with tech support and R&D on this, but they have not mentioned the configuration manager.
I have determined that I can successfully start MAX only when I am connected to my office network. When I am away from the network, I get the error. All the other things I had tried before (re-installing drivers, recovering the data folder, etc) were not the reason MAX worked, it was because I was re-connected to my office network at the time.
I currently have MAX running as I started it upon leaving the office and don't want to close it until I can get back to the office again (I have a remote office). Currently the configuration mangaer is running. I have not tried changing any settings. What might I change?
My feeling is that it is searching for something that happens to be on the network. Since I am not connected, it can't find it, and doesn't start.
04-30-2008 02:25 PM
We found the solution. My .iak file was located on a network-syncronized folder. I got the error whenever I was disconnected from that network. As soon as I reconnected to the network MAX was able to load. So, i moved the .iak to a folder on my c drive, instead of the network folder. NI R&D is fixing the problem in the next FieldPoint release.