Hello Drew,
Are you Daisy Chaining you 4 SCXI chasses? If so, AI0 - AI3 will be reserved for your 4 chasses. You should be able to wire the counter output to AI4 of your DAQ device.
Let me know if you ahve any further questions.
Regards,
Sean C.
07-18-2005 05:17 PM
07-19-2005 11:46 AM
Hi Drew,
The SCXI Chasses each multiplex all signals to a single differential line. This means the AI0 - AI3 AND AI8 - AI11 are being used by the SCXI Chasses.
Also, if you are using AI Sense as the reference voltage for your AI channels, make sure that you are measuring in NRSE mode.
Try using AI4 and see if the problem still occurs.
Regards,
Sean C.
07-19-2005 01:41 PM
07-20-2005 02:04 PM
Hi Drew,
Can you verify that you can read correctly from this channel without the SCXI Chasses connected? Try connecting your power supply directly to AI 4 on the PCI-6052E (pins 28 and 27) and read it in RSE Mode. If this works, I would start looking for bent pins on all cables, SCXI-1346s, and your DAQ device.
Also, have a look at the following Knowledge Base: http://ae.natinst.com/operations/ae/public.nsf/web/searchinternal/09c41ee1e479893286256b02007cc76f?O... .
How are you connecting your SCXI-1302? I am assuming that you are using the SCXI-1180. If so, what are you connecting the 1180 to? Is it a SCXI-1346, or a SCXI-1349? You cannot use a SCXI-1349 to connect your 1180 if it is also connected to a parallel capable module (such as the 1520).
07-20-2005 05:28 PM
07-21-2005 01:05 PM
07-21-2005 03:37 PM
07-22-2005 01:47 PM
Hi Drew,
Can you reproduce this behavior from a test panel in MAX? Try setting up the pulse train generation from MAX and wire the counter output to an analog input. Then view that input channel in MAX. Does it still vary from -2.5 to 2.5 Volts? Is it a square wave?
If you observe the same behavior in MAX, try performing the same test without the SCXI hardware connected. Does this solve the problem?