Keithley 6517B driver error

Solved!
Go to solution

Hello,

I'm working with the Keithley 6517b electrometer and wanted to use the LabView driver. However while testing the example "Acquire measurement" the error queue produced the attached errors.

Doing measurements with a selfmade LabView VI works..

Does somebody have an idea?

Thanks a lot.

Regards,

Jonas

0 Kudos
Message 1 of 4
(3,514 Views)
Solution
Accepted by topic author JonasO

Hi Jonas,

 

Does somebody have an idea?

Yes: it's your fault! 😄

You are using a German (?) Windows using German regional settings with a comma as decimal separator.

Almost all device drivers from NI rely on using a point as decimal separator - as well as 99.9% of all external DAQ devices I have seen so far.

So you get an error when you try to scan strings expecting the "wrong" decimal separator…

 

Solutions:

  1. Set your Windows regional settings to use English number formats with points as decimal separator.
  2. Write your own device driver handling such regional setting issues…
Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 2 of 4
(3,508 Views)

Thanks a lot..

That didn't cross my mind. Dumb 😄

Regards

Jonas

0 Kudos
Message 3 of 4
(3,493 Views)

Hi JonasO,

 

on top of GerdW's answer, this may be useful as well: Localization Codes in LabVIEW.

 

Regards,

 

 