Good suggestions Space_Flight,
Here are some additional suggestions:
In your application you are performing buffered period measurement. This uses implicit timing which does not latch based upon a clock, it latches based upon the edge of your signal you measure. For low frequency 1 counter measurement, the counter will begin counting edges of the timebase prior to the first edge seen on the counter's gate. For this reason, the first measurement will only be a partial count and will resort in a short period measurement for the first measurement. All other measurements will be good.
Also, you are synchronizing by using a shared clock. This will ensure that measurements are taken at the same time but does not guarantee that you start at the same clock edge. You may get one or more samples on the task which you start first before the other task begins. To avoid this you must use a DAQmx trigger function or use the AI sample clock for the DI task as shown in the Multi-Function-Sync AI-Read Dig Chan.vi example (Help>>Find Examples). Space_Flight's suggestion of using the AI/Sample Clock will work even if the sample clock for your analog channel is an external source on a PFI line.
Hope this helps,
Jennifer O.
Applications Engineer
National Instruments
12-06-2006 02:06 AM - edited 12-06-2006 02:06 AM
Message Edited by Greg H on 12-06-2006 02:09 AM
Message Edited by Greg H on 12-06-2006 02:10 AM
12-06-2006 09:21 PM
12-06-2006 11:54 PM
12-07-2006 06:50 PM - edited 12-07-2006 06:50 PM
Hi Greg,
I'm still not quite sure where the digital task fits in. You refer to position measurement and determining speed. This would generally be done with either a count events, frequency measurement, pulse-width measurement or position measurement task. This would be a counter task and not a digital task. If we timed a digital task off of the rising edge of the encoder, we would simply be ensuring a high measurement every time, providing no useful informaton.
Secondly, the timestamp is generated by software. This is explained by this KnowledgeBase: Why Is the Waveform Timing Information Returned by NI-DAQmx Incorrect? This will not be an accurate way to determine speed if the motor or encoder signal is not constant.
I would look at using the DAQ board for timing (internal timing). You can set up two tasks. One would be analog to measure the voltage and current. The other task would be a buffered counter measurement. You would clock it from the ai sample clock. This would give you a constant acquisition rate on the analog channel and thus give you more accurate timestamps. You would start the counter task first and then start the analog input. No triggering is required to synchronize the measurements and the measurement would be independent of whether the motor is started or not. The counter will begin counting immediately, however you can add an arm-start trigger to keep it from counting until the analog sample clock begins.
You may also be interested in reading this post. Space Flight posted this picture, which will guide you in using the above option. You will want to add the following code to the counter task to control the start of the counting:
Regards,
Jennifer O.
Message Edited by Jennifer O on 12-07-2006 06:52 PM
12-21-2006 10:48 PM
12-22-2006 09:17 AM
I am using an M series card (PCI 6259) and labview version 7.1.
I am using a 12-bit absolute encoder which comes in on port0. I have added some harware logic so that a short pulse is created each time the encoder changes. That pulse is wired to PFI0.
The goal is to take current and voltage readings (analog) and encoder reading (digital) along with a time stamp (to determine speed later) each time the encoder changes (PFI0).
Here's what I'd do:
1. Configure DI for "Change Detection" based sampling to eliminate the need for the external circuit. Specify sensitivity to both rising and falling edges on all your 12 bits. You'll need DAQmx 8.0 or later for change detection support on M-series boards.
2. When calling DAQmx Timing for the AI task and the Counter task, specify "ChangeDetectionEvent" as the source for the sample clock. You should probably start those 2 tasks before starting the DI task -- that way they'll all be ready to take a sample on the very first DI change.
3. To measure timestamps with your counter, choose the Edge Counting mode. The edges you count should be configured to use one of the board's internal timebases, like 100 kHz or 20 MHz. I don't recall for sure, but you *might* need to use a DAQmx Channel property node to specify one of these timebases as the source for the edges to count. What you'll get: a cumulative count of # cycles of a clock with a known period, i.e., a timestamp.
-Kevin P.