Help to program a Ni 6009

Capture1.JPGCapture2.JPGCapture3.JPGHello that put in the input signal range of the DAQ assistant with the voltage and current knowing that my ni 6009 will be between a solar panel of 17.6v and 2.65a and a oscilloscope of caliber 1/200 and a 100mv / a caliber amperemetric clamp, thank you very much

 

 

Bonjour, que mettre dans les donnée gamme de signal d'entrée de l'assistant DAQ avec la tension et le courant sachant que mon ni 6009 sera entre un panneau solaire de 17.6v et 2.65a et un oscilloscope de calibre 1/200 et une pince amperemétrique de calibre 100mv/a, merci beaucoup

0 Kudos
Message 1 of 4
(2,581 Views)

Hello,

 

Perhaps is is a language barrier and a failure in your translation to English, but what is your question?

0 Kudos
Message 2 of 4
(2,546 Views)

I would like to know what I have to put in the round screenshots, with a 17.2volt solar panel and 2.65 ampere, with a 1/200 caliber oscilloscope and a 100milivolt / ampere current clamp meter

0 Kudos
Message 3 of 4
(2,529 Views)

For volts, put in +20 and -20.    The NI-6009 has a +/- 20V range, and you are trying to read 17.5 V.  Pay attention to that footnote about the inputs needing to be within 10V of ground.

 

I don't know what you mean by a 1/200 caliber oscilloscope.

 

For 100 mV/amp current meter, since you are trying to read 2.65 A, that means you'll be getting 0.265 V.  Your lowest voltage range for the NI-6009 is +/- 1V, so put in +1 and -1.