06-13-2022 05:59 AM
Hello all,
When I acquire data using cDAQ NI 9215 and chassis 9171, my data is getting tripped/hatched by the lower side ( I have attached a file for your reference). And this happens with some channels at different times while other channels give good output. Kindly, help with the problem with this. What can be done to rectify this problem?
Figure 1: tripped/hatched signal
Figure 2: tripped/hatched signal
Figure 3: Estimated result.
Note: I also checked with NI MAX Test Panel, I am getting similar results.
06-13-2022 06:52 AM
Are these signals floating? Or the grounds are not referenced to each other? That is what this looks like.
You might want to have a look at this article: Field Wiring and Noise Considerations for Analog Signals
06-13-2022 07:51 AM
Most likely (IMHO) a missing Input Bias current path. Solutions in the link crossrulz provided... (one or two 1Meg resistors from the anlag - input (or both) to AGND usually solve that.
06-13-2022 11:00 AM
I have noticed similar results when the vertical range is not correct.
06-13-2022 01:46 PM - edited 06-13-2022 01:47 PM
I am using NI 9215 and Current Transformer sensor connected vai BNC cable.
06-13-2022 01:48 PM
Can you tell me how to correct this?
06-13-2022 04:38 PM
Your sensor is likely floating. This means it needs something to pull it towards ground. Added 1MOhm resistors on each side of your transformer to AIGND should do it.
06-14-2022 01:44 AM
Thankyou for your answer. Can you guide me how to ground this using resistor. I am using current transformer and a BNC cable for this.
06-14-2022 03:24 AM - edited 06-14-2022 03:31 AM
Which type of 9215 do you have? The BNC input version should have a 100kOhm Resistor to AGND/COM already ...
Which version are you using?
For the screw or spring version try a 1M Ohm resistor between AI- and COM terminal
If you use the 9215 BNC version and you run into that problem, please provide a schematic of all your cabling. Do you have a handheld multimeter (DMM) around? can you check the voltages at the BNC shield .. between the shields and to PE/Ground ?
I don't have a 9215 (BNC) around, but I assume you should measue a resistance of 200kOhm with the DMM from one BNC connector shield to the next channel BNC connector shield. (Just in case one 100k is blown in history)
06-16-2022 03:21 AM
I am using 9215 (BNC).