1. Probe the output of the GPIB read function and look at the string returned. The instrument itself may not be returning any more significant digits. It may be a limitation of the intrument or require a command to be sent. Looking at the string will tell you whether it's the VI or the instrument.
2. The instrument will not do a service request until the measurment is complete. With a 10 second gate time, expect at least 10 seconds for RQS. It may be longer.
3. I wonder if the status register in the instrument needs to be cleared. There's a tutorial here on SRQ servicing.