Agilent E4980A configure correction

Hi All,

 

I am working on automation VI for Agilent E4980A. 
I started from old sw compatible with the old LCR meter Agilent 428X. 

I am trying to configure correction in the VI flow but I continue to have the following error:
 

Error -1074000000 occurred at Agilent E4980.lvlib:Error Query.vi

Possible reason(s):

Instrument reports:
+71,"Scanner I/F not installed"
+0,"No error"

Mdr20_0-1677777325499.png

below the old vi with the basic correction setup:

Mdr20_0-1677777420747.png

They look the same.

 

Do you have any idea of how to solve it?

Thanks in advance.

 

 

0 Kudos
Message 1 of 3
(1,135 Views)

The seems pretty clear that the instrument say something is not right. I would first check what Error +71,"Scanner I/F not installed" means in the User manual of E4980A and see what troubleshooting steps the manual recommends.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 3
(1,120 Views)

Hi Santo_13,

 

I checked also on user manual:

 

When Option 301 (Scanner Interface) is not installed and you attempt to switch to multiple
correction mode, a Scanner I/F not installed error message appears in the system message
area.

 

The fact is that I am trying to have a single mode, that's why.

Do you know any solution?

0 Kudos
Message 3 of 3