Hi, Agilent Users,
 
I'm using 34970A with 20channel multiplexer module and try to scan the voltage signal and save the data file using LABVIEW 8.2.
Communication with Labview and the 34970A is made via USB-GPIB and the example code that I downloaded from NI instrument driver sites works fine
 
But, I have difficulties in scanning continuously for a certain time, let's say, 2 minutes with 10 ms. The example code, "The advanced scan example.vi" is for only one scan.
So,I wanted to modify the program being able to scan the channel with the configuration explained above. I set the interval time,  max trigger time and any other parameter that might affect the scanning function but,  I couldn't get results that I wanted.
Any idea or comment are welcome and if you have a code for that purpose, it would be appreciated if you can share the code.
 
Thank you very much.
 
Wonseok Yoon
 
 
 
 
0 Kudos
Message 1 of 17
(7,821 Views)
How are you trying to make it run continuously? Did you put the VIs in a while loop? What sort of behavior were you getting?
Rasheel
0 Kudos
Message 2 of 17
(7,801 Views)

You could create a loop and call the single scan VI example as a sub-vi within the loop.

If you tried that and had problems, then you could post your code.

RayR

0 Kudos
Message 3 of 17
(7,797 Views)

Thanks, guys,

I appreciated your comments.

I tried to scan channels with "Immediate" source with scan to scan interval of 10ms, # of trigger (300), but I realized that was wrong.

Today, I figured that out  using "Timer" source with configuration above and the VI can scan the channel that I set without using "While" or "For" loop.

However, I got another problem. I want to get more than 300 scanned data per channel, but the data shows in the front panel is only around 113 data, not 300 data.

I don't know why the VI can not read 300 data.

The attached are front panel and block diagram of my VI.

Thank you very much.

 

Wonseok Yoon

Download All
0 Kudos
Message 4 of 17
(7,789 Views)

Are you requesting enough data to read from the instrument?  I would have to check the actual commands, but although you tell it to do 300 samples, the size of the data to read may be a larger number.

Question:  Are you trying to remove the commas in that loop?? 

RayR

0 Kudos
Message 5 of 17
(7,774 Views)
If you want to remove the commas, try the example below using the Search & Replace string function.
 
RayR


Message Edited by JoeLabView on 06-19-2008 04:43 PM
0 Kudos
Message 6 of 17
(7,765 Views)

Hi, Thank you for the suggestion.

What do you mean by "requesting enough data to read from the instrument?" and " the size of the data to read may be a larger number."

 

I didn't remove the comma in the " while" loop.

The weird thing is that if I use, let's say, # of trigger of 10, then the VI exactly can read 10 data of each channel. but If I set the number as the value more than 100,

then, the same thing happens.

I'll post the results tomorrow after I try yours.

 

Regrads,

 

Wonseok Yoon

 



Message Edited by swsyoon on 06-19-2008 03:52 PM
0 Kudos
Message 7 of 17
(7,758 Views)
Hi Wonseok Yoon,
 
So if I understand correctly, you are getting reasonable results up to a value for # of Triggers of 99.  When you go to 100 triggers and beyond (with a target of 300), then you do not get all your data.  Is that correct?
 
So for 10 channels with a trigger of 99, you are looking for 990 points of data.  For 300 triggers, you are expecting 3000 data points.
 
In your VI, the HP34970 READ...  does it have a size configuration for the number of data points to read?  I do not have that driver installed, so I do not know what configuration is available.  You can select from the HELP menu, display Context Help, or click on the question mark at the top right corner of the block diagram window to display Context Help.  Place your mouse cursor on top of the HP34970 READ and look at the description.  (also posting the picture would help 😉 )  It may have a default maximum number of bytes to read which is equivalent to 99 data points.
 
I just looked more closely to your While Loop.  You're converting the text to an array of strings, delimiting it by a comma..
Okay..  I'll post another example of a simpler way to do this.
 
RayR
 


Message Edited by JoeLabView on 06-19-2008 06:13 PM
0 Kudos
Message 8 of 17
(7,735 Views)
Hi Wonseok Yoon,
 
The image below is the example.  You do not need a loop to convert the text to an array.  You can replace the While Loop with a "Spreadsheer String To Array" vi.   It is located in the strings functions. 
 
You simply wire the string from the HP34970 READ to the spreadsheet string input, place a blank string at format string, place a comma for the delimiter (on top), and very importantly, wire a sting array constant to the array type (bottom).
 
RayR
 


Message Edited by JoeLabView on 06-19-2008 06:25 PM
0 Kudos
Message 9 of 17
(7,729 Views)
Another question:  did you actually want to have a 2-D array for Read?
 
Can you describe that portion of the code?
 
RayR
0 Kudos
Message 10 of 17
(7,719 Views)