09-20-2012 06:21 PM
All,
I have an error when trying to use the Analog Devices ADXL346 with the NI USB-8451 over SPI. I recieve the following error:
"Error -301713 occurred at Property Node (arg 1) in SPI_Simple_forum.vi
Possible reason(s):
NI-845x: An input parameter, or combination of parameters, is invalid. An example of this error may be setting an invalid baud rate or enabling I2C ACK polling while using 10-bit addressing."
The interesting thing about this is it only occurs when the I/O voltage is anything but 3.3V. When it is 3.3V I do not get the error. However, the I/O voltage of the device is set at 2.5V so I believe I need to set the 8451 I/O voltage to 2.5V and thus get the mentioned error. Because the error only appears when I change the voltage, I can conclude it doesn't like the voltage selection but I don't know why. Any insight? Thanks.
My VI, and datasheet are attached.
Solved! Go to Solution.
09-21-2012 04:18 PM
Hey,
Sorry, but I'm pretty sure you're getting the error because the USB-8451 is not capable of variable voltage levels. If you look at the NI-845x Hardware and Software Manual, in Appendix A, it does not mention anything about being able to change the voltage levels on the USB-8451. If you scroll down a little to the USB-8452 section, it does. If you have any more questions about this, please feel free to ask!
Best,
09-21-2012 04:27 PM
Thanks Jeff. I see that. I basically copied the NI example for reading and trimmed it down for this chip. I guess the example is more geared to the 8452. So does this mean the 8451 cannot work with this accel since the supply voltage is 2.5V and a supply voltage of 3.3V would exceed the maximum for this device?
09-24-2012 02:29 PM
I didn't look at the spec sheet for the ADXL346, but I wouldn't use an out of spec voltage with it.
11-05-2012 06:00 PM
I have acquired an 8452 and with that set to 2.5V the ADXL346 works perfectly