to control Agilent 4156C via LAN interface but get a error display“Cloud not open a VISA section to TCP/IP...,VISA error code OXBFFF0079"

att as the tittle,my PC OS is windows7, labview 2009 and visa 5.02 are installed on, there also is a exchange board, when connected with Agilent 4156C, there is a error display, and it means ”The protocol specified is invalid“ according to the reference. what's the problem? thank you!

0 Kudos
Message 1 of 6
(4,181 Views)

What protocol did you actually specify? Did you select the resource name from the pull-down menu? Did you create the resource in MAX?

0 Kudos
Message 2 of 6
(4,160 Views)

when I create the VISA TCP/IP resource and manually specify the address information of the LAN instrument in MAX, the error display acts.

0 Kudos
Message 3 of 6
(4,135 Views)

You specified Raw or VXI-11? Try swapping the resource type. Also, make sure you can ping the instrument.

0 Kudos
Message 4 of 6
(4,124 Views)

Thanks,Dennis Knutson. I specified the VXI-11/lan. when I pinged the instrument IP(192.168.0.9), there were  no lost packages and the time showed 1ms. I only installed labview2009 and visa 5.02 on my PC, can I create the TCP/IP resource in MAX? should I intsall other drivers of Agilent instrument?

the attachment is the screen shot  I geted when I operated.

0 Kudos
Message 5 of 6
(4,106 Views)

可以参考一下这个链接中的方法,看是不是有效果。

Haifeng Xu

NISH AE