error -1073807339

i using wireless modem to connect with logger to get data..
i use the serial communication  and i able to send the request out for example the command is like tis

?00001*RR\r.. but wen i didnt get the reply from the logger

if i use the hyperterminal i able to get the responce..

 

this error is occurs wen i request -1073807339

VISA Read in Serial Port Read.vi  why this occurs

 

and the byte count always returns 0 when requested..

 

why this occurs?

 

plz help

0 Kudos
Message 1 of 3
(3,086 Views)
This error is due to a timeout.  You should use the VISA Set Timeout VI to increase the timeout value.  This information can be found in KB 3JOF9K6K.
Ben

National Instruments
Certified LabVIEW Associate Developer
Certified TestStand Developer
0 Kudos
Message 2 of 3
(3,067 Views)

But, the reason the timeoout occurs may be because you are not sending the termination character correctly. You will only send a '\' code if the string control/constant is set for '\' Code Display. Right click on the string control/constant and see if that is checked. You can set the timeout to infinitity and that won't help if you are not sending the correct command.

 

 

0 Kudos
Message 3 of 3