Hi Joe,

In order to help you better with this issue, I would like to ask that you contact a support engineer via email, so that we can communicate with you directly.  Please let them know that you have already done some troubleshooting from the forum, and ask them to contact me regarding this issue immediately.  You may use the link below to contact a support engineer by email:

Request Support

Thanks,

Jason S.
GPIB Software
National Instruments


0 Kudos
Message 11 of 12
(1,414 Views)
Hello Mr. Ohle,
in the past we worked on a case where an other customer had problems similar related to yours where our older GPIB boards work with older Stanford Research Systems instruments, but our newer GPIB boards failed; at least not without some modification in the connection between the board and the instrument.

The workaround was to place a 270 pF capacitor between pin 10 (ground) and pin 17 (DAC) of the instrument's GPIB connector/cable.

Maybe that will help in your case too, but please keep in mind that NI is not reliable of damages which may occur at your device, GPIB board or computer based on making this kind of modification.
.
Kind Regards,
Heiner Illig
NI Germany