Janet,
Typically when the installations fails the sequential verification, it is due to either a resource conflict or a device being connected to the board during the test. You can click on the failed interface to get a brief explination of what may have caused the error.
There are instances reported using the AT-GPIB/TNT+ board and Win 98 inwhich this error was reported even though there were no problems with the board's setup. Since you are communicating properly with the board through IBIC your board is functioning properly, and I would not be concerned with the error message.
As for your instrument, this is likely due to your instruments non-compliance to 488.2 standards. Whenever the Measurement & Automation Explorer does a "Scan for Instruments", it sends out the stri
ng "*idn?" (which is an IEEE 488.2 identification query string) to all of the devices on the bus. If you have older, non-IEEE 488.2 compliant instruments on your GPIB, then they will not recognize this string as a valid command. This will cause "Scan for Instruments" to fail.
Best Regards,
Chris D