Hello,
Im using a Chroma Power supply 62000P via USB cable (USB A to USB B), I have installed Keythey IO Suite 18.0 and NI MAX 17.0, NI VISA 15.0 and Labview 2012, I created a program to control the power supply and perform a sequence, all was working good for about 2 months, now I cannot connect, I get the error: VISA: (Hex 0xBFFF003A) Unable to start operation because setup is invalid (due to attributes being set to an inconsistent state).
When using the VISA Test Panel and sending a ID command i get the same error.
I have tried restarting Computer, plugging the instrument again and reinstalling Keysight IO Suite, any ideas?
Thank you
Not sure why you need Keysight IO Suite installed.
Did you ever figure this out. I installed a bunch of NI updates yesterday to LabVIEW 18 and now I can't talk to any of my Keysight devices. Getting this same error. I'd rather not install the Keysight I/O suite, but I'm out of ideas.
@TobyD ha escrito:
Did you ever figure this out. I installed a bunch of NI updates yesterday to LabVIEW 18 and now I can't talk to any of my Keysight devices. Getting this same error. I'd rather not install the Keysight I/O suite, but I'm out of ideas.
No, I had to uninstall everything, from Labview to VISA and keysight drivers of another device I had, then I installed Labview and Visa and the keysight drivers and it worked.
That's what I was afraid of. There goes my Monday! Thanks for the update.