Serial Communication with TDK-Lambda
Hello, I'm trying to control the genesys TDK-Lambda 750W power supply through serial communication(Through hyperterminal). I have connected my power supply with the PC using a RS 232 cable with DB25 and Rj-45 connectors. THe power supplu has the following settings:
Bits per second:9600(Default)
Parity :None
Address:06(Default)
For serial communication the power supply is setted to use RS232 in the rear panel.
I have verified the address, the baud-rate, the switches on the rear of the instrument, but it appears all ok.
I have verified the COM in the PC, too..but it doesn't talk with the tdk lambda. when I send "ADR 06" to the power source through hyperterminal it should say "OK", but I read only
ADR 06
ADR 06
I don't know whether the communication port on the power supply is enabled or not.could any please help. Thank you
Message 1 of 9
(15,817 Views)
Neither hyperterminal or the instrument are NI products so I'm not sure why you are posting here. The vendor should have a support page.
That being said, double check the command and the termination character(s).
Message 2 of 9
(15,809 Views)
You are the one sitting in front of the instrument. How are we to know?
Message 3 of 9
(15,802 Views)
Baud rate needs to be set to 19200.
Have fun.
Message 4 of 9
(15,550 Views)
mraya001, were you able to get this to work?  I have the same Lambda power supply with the Lambda-supplied RJ45 to DB9 RS-232 cable and I get no response from the power supply whatsoever, just a timeout.
The port on the supply is enabled as far, as I can tell from the documentation.  The PC's serial port is functioning because I can use the NI-MAX VISA Test Panel to communicate with two other devices that are presently sitting on my desk.
Unless I missed it, the documentation does not say anything about having to send special termination characters, but it wouldn't surprize me if that's what it is.
Message 5 of 9
(15,303 Views)
Ok, I got this to work using the VISA Test Panel in NI-MAX.  It was the termination string which Dennis alluded to in a previous post.
Make sure you uncheck Send End On Writes.  The power supply requires only a carriage return after a command.  Nothing more, nothing less.  Oddly, it will not work if you check Enable Termination Character.  Once I unchecked those and manually added the \r to the end of my command strings, everything worked.  Also note that the supply will not respond if it doesn't get a ADR\s06\r as the first command in a "session" (the \s and \r are just visual representatives of a space and a carriage return respectively.  06 is the factory default address setting).
(edited for spelling)
Message 6 of 9
(15,296 Views)
Friend, I tried the way you mentioned, but I still have no communication with the Power Supply.
Could you share the pinout you used on RS232 cable with DB9 connector to RJ45 connector?
I made following the table described in page 51 from TDK-Lambda User Manual:
DB-9       RJ45
2 RX       7 TX
3 TX       8 RX
5 SG       1 SG
Note: The DIP Switch "SW1" are all in the DOWN position!
Message 7 of 9
(3,956 Views)
I had to make an account on this site, just in the hope that I might save someone else the headache that I've been through trying to figure this out.
After banging my head against the wall for days, trying every type of variation I could come up with, I finally went back to examining the cable pinout(I made a cable instead of buying one).
To keep this short I'll just attach images that explain why I had made my first cable improperly.
RJ45 Jack Pinout:
RJ45 Jack PinoutRJ45 Jack PinoutTDK Lamda RJ45 Jack Pinout:TDK Lambda RJ45 Jack PinoutTDK Lambda RJ45 Jack Pinout
 I hope this small post saves someone else the dread of getting no response for hours and hours.
Message 8 of 9
(1,318 Views)
Exactly. The company didn't use the standard RJ-45 pinout. They ended up inverting the pin order.
It also took me several days to identify this subtle "problem."
Message 9 of 9
(1,302 Views)
