Using NI MAX, going to the RS232 interface, validating the interface OK.
Opening the VISA TEST PANEL sending commands OK.
Receiving the data, OK after the specified timeout. But because it needs to time-out before it sends the data, I get in the RETURN STATUS the following:BFFF0015
With the VISA Configure Serial Port, do you have the Enable Termination Char unwired or have a True wired to it? If either, that tells VISA to expect a termination character with every read. The VISA Read will not terminate until the number of bytes specified is read, the termination character is detected, or the timeout is reached. The default termination character is a LF character (\n or xA). Check if this is what is being sent if you want to stop the read with a term character. If no termination character is being sent, use VISA Bytes at Serial Port to dtermine exactly how many bytes are available for reading and wire this value to the byte count input of VISA Read.