08-15-2016 08:14 PM
Hello all,
I have different instruments connected trough GPIB, including a Lakeshore 335, and everything worked fine. The problem came when we recently bought an old model (330). The 'scan for instruments' does not find it, even when using all the same cables as the model that does work. The manual of this old model http://www.lakeshore.com/ObsoleteAndResearchDocs/330_Manual.pdf is quite outdated and I only find information on how to connect it to LabVIEW, which I don't use. I tried changing the MAX GPIB interface properties to match the ones specified for the old version in the manual, but I am not sure if the EOS byte 0AH correlates with number 10. I tried different numbers and nothing changes. Does anyone have any ideas on why it doesn't find it? Anything will be appreciated.
Thank you,
Laia
08-16-2016 12:59 PM
Hi NterLab,
The EOS byte 0x0A (which is the same as 0AH) is the Hex value for a Linefeed Constant. Can you describe where the number 10 is coming from?
I’ve attached a link to some common troubleshooting steps. Let me know what on this list you’ve tried so far.
http://digital.ni.com/public.nsf/allkb/72FAD7156202DDF9862568900048F006?OpenDocument
Can you attach a screen shot of where you’re changing the MAX GPIB interface properties? This might help lead to a change that needs to be made.
Do you have any other devices connected when troubleshooting this device?
08-16-2016 02:58 PM
Dear Matt,
Thank you so much for you reply. I attach my MAX screenshot. The byte I though 0 was 00H but if 0 is 0AH then I change it to 0, but it still does not work. I have 5 other devices connected, all work fine. The Lakeshore 330 is the only one that does not appear. Its address is 12, which no other instrument has. I let the device off the whole night, but nothing changed. I also checked the NiVisaTulip.dll
This is what I typed in interactive control and the error it gives me
ibfind gpib12
[8000] ( err )
error: EDVR (0xe0140025)
I tried to use ipdev instead but I do not know most of the parameters it asks; except for the following (extracted from the manual)
Primary GPIB Address ........→12
Secondary GPIB Address ...... NONE
Timeout setting ............. 10sec
Serial Poll Timeout ......... 1sec
Terminate Read on EOS ....... Yes
Set EOI with EOS on Writes .. Yes
Type of compare on EOS ...... 7-Bit
EOS byte .................... 0Ah
Send EOI at end of Write .... Yes
Enable Repeat Addressing .... Yes
If you can please tell me what should I type after ipdev, that would be very helpful.
Thank you again for your help,
Laia.
08-16-2016 05:04 PM
Screenshot here
08-17-2016 05:30 PM
From my understanding of ibfind gpib, this initiates communication with the GPIB controller, not the device you’re trying to find. I’d recommend following the steps in the link below to see if you can find lakeshore 330.
08-18-2016 04:18 PM
Dear Matt,
I have followed all the steps and nothing finds this instrument. I have tried different addresses, cables, and NI boards too, in all the cases, after using the interactive control, MAX sees all the instruments but the LakeShore 330. Do you have any other suggestion? I think that the NI integrated circuit may be broken but I don't know how to test it.
Thank you,
Laia.
08-19-2016 08:08 AM
Hi Laia,
I’m not sure what you’re referring to when you say the NI integrated chip might be broken. Can you provide some context for this?
Have you used the front panel on the Lakeshore to verify the address and the termination character? Also, you said this was an old model, so have you tested it on its own to make sure it works?
Once you’ve verified the address, termination character, and operation of the Lakeshore, I remove all other devices and send some known commands to the device, something like MODE?
08-19-2016 04:53 PM
Hi Matt,
I mean the NI chip inside the instrument. I think that something is wrong with the instrument. It may be the GPIB port that is broken, although everything looks normal and I don't know how to check it.
The instrument works properly when selecting the settings in the front panel, it controls the temperature well. Also, I tried different addresses (12, 15,18...), using the front panel, and I tried all the possible terminations. The termination options are
CrLf Carriage Return, Line Feed
LfCr Line Feed, Carriage Return
Lf Line Feed
End No Terminator (EOI line set with last data byte if active)
Nothing works.
If I try to connect using the interactive control (ibdev 0 12 0 13 1 0x140A) and ask it the address using ibwrt "*idn?" it gives me an error (ENOL). Also, any other command like ibwrt "*CLS" or ibwrt "*MODE" give me the same error.
Laia.
08-19-2016 08:22 PM
Have you thought about using the serial port?
Curt
08-20-2016 11:14 AM
Hi Curt,
I don't have an adapter for the serial port, but you're right I should try that first...I am going to go buy one. Thank you!
Laia.