How big is the message queue in GPIB?
I'm wondering how big is the message queue in devices woking with GPIB interface?
Can I for example ask the device to deliver some results from the SpecAn first and put them to the output queue for example:
(pseudocode)
Put Marker 1 value to the message queue
then read them back to the computer.
Get Marker 1 value from the message queue
Instead of doing like below:
Can the queue holds a lot of results coming from a few measurement requests to be read back as the block?
Is it vary from one device to another or is itdefinied by the GPIB standard?
(3,515 Views)
This is depending on device's design and, whether if the Put and Get action is forming a pair (such as Query and Response) or not.
If the device is compliant with IEEE488.2 or SCPI standard, and if the Put and Get are a pair of query/response, there is an obvious rule.
IEEE4882 and SCPI specifies a rule that a command resets the pending response cue that was generated by the previous query.
Example:
MARKER:VALUE 1
MARKER:VALUE?
(you can read an answer --- 1)
But,
MARKER:VALUE 1
MARKER:VALUE?
MARKER:VALUE 2
(you can not read any answer  because the old query was cancelled)
(3,512 Views)
Makoto,
Many thanks for the info. t was very helpful. Especially this statement:
(...) IEEE4882 and SCPI specifies a rule that a command resets the pending response cue that was generated by the previous query. (...)
Any chance you can reference the document in which these feature is described?
Cheers.
(3,506 Views)
