If you are using an Agilent GPIB controller (why?) with LabVIEW and not an NI one, then you should look at http://digital.ni.com/public.nsf/websearch/F7C187DBF09EBE1186256F550065BD32?OpenDocument.
If you can't communicate with the instrument with the Agilent utility either, then there is something wrong with the Agilent controller, with the Agilent drivers, or the instrument. Unless you replaced the Agilent drivers with ones from NI, then I'm not sure that this is the place you should be asking questions about your problems.
05-08-2009 10:09 AM
I had the same problem before, but I solve that by including a minimum waiting time of 5 seconds. This is because the 4145B is so old that it takes some time in providing data due that it works with a start-up diskette. Anyway I made an apllication for measuring the IV-Characteristics of a device attached to SMU1 and SMU3 terminals of the 4145B analyzer. SMU1 is a voltage source with a compliace of 100mA. This is a LabView 8.5 application partially inspired in ni drivers for 4145B posted by Natali and others. I hope it works smoothly and enjoy it.
Henry Mendez (Colombia).