HOW to change myHigh speed 488 to low speed 488 ?
The LED of my  GPIB-USB  card  is green  a few month ago ,but,resently it changed to Orange! Then it can't recognize the instrument! What should I do ?
(3,131 Views)
Check your set-up!!!  The most common cause for this type of behavior is that your configuration changed to the point where you've violated a bus phyisical requirement.
 Quoting fro NI-488.2 help file----
"Configuration Requirements
"To achieve the high data transfer rate that the GPIB was designed for, you must limit the number of devices on the bus and the physical distance between devices. The following restrictions are typical:
  • A maximum separation of 4 m between any two devices and an average separation of 2 m over the entire bus.
  • A maximum total cable length of 20 m.
  • A maximum of 15 devices connected to each bus, with at least two-thirds powered on.
For high-speed operation, the following restrictions apply:
  • All devices in the system must be powered on.
  • Cable lengths must be as short as possible with up to a maximum of 15 m of cable for each system.
  • There must be at least one equivalent device load per meter of cable.
If you want to exceed these limitations, you can use a bus extender to increase the cable length or a bus expander to increase the number of device loads. You can order bus extenders and expanders from National Instruments."
"Should be" isn't "Is" -Jay
(3,114 Views)