GPIB to Serial Converter - problem with receiving RS232 and seeing it on GPIB (Service Request 7198845)

Situation:
I'm trying to use a GPIB to Serial Converter to control a temperatuer chamber controller.
I'd like to use the interface in D mode, so that it is supposed to be transparent,
just a GPIB to RS232 converter with a fixed setup for baud rate etc.
Why:
The PXI chassis I have is practically full and all my regular COM ports are used.
The chamber controller has only RS232 interface.
 
Easy way you can reproduce the problem:
In order to illustrate the problem I've used it also in G mode, where the commands to the interface and the data use separate GPIB addresses.

I have placed the device in G mode, with address 16 for the interface and 17 for the data (serial) part.
Address 16 replies to the "id\r" query (from MAX) as expected.
 
As the knowledge base suggests, I have connected the TXD and RCV pin together for a loopback.
 
Any read from address 17 (using a query or a Write and then a Read) gets EOAB errors.
I have tried to use the various termination characters (\r, \n or a combination) - no deal.

When I use the interface in D mode and connect to the actual serial device (instead of the loopback),
I can see that my WRITE works (commands are executed) but my READ does not -
even when I see a reply with a scope, the NI Spy does not show any data coming back).
0 Kudos
Message 1 of 5
(3,849 Views)
Hello,

Thank you for contacting National Instruments. It looks like someone is already working on the phone support issue (7198845) and will get back to you shortly.

Regards,
Simran K
National Instruments
Applications Engineer
0 Kudos
Message 2 of 5
(3,819 Views)
Thanks for the answer.
 
However, so far there is no resolution and I was wondering if anybody else has encountered the problem and maybe even found a solution.
Is it possible that I'm the only one using the product?
 
Thanks
 


Message Edited by Eagle Eye on 06-19-2008 06:18 PM
0 Kudos
Message 3 of 5
(3,766 Views)
The solution to this problem boils down to termination characters. The complicating factor is that we are using both the convert box borrow a few settings from both the GPIB and Serial communications.

Solution for G-mode:

The EOS setting must be set on the gpib-rs232 converter box by sending an  the command "eos [X[B]eoschar]<CR>, where X or B is the termination mode

Description Letter
XEOS—Set EOI* with EOS when sending data from serial device X
BIN—compare all 8 bits of EOS byte rather than low 7 bits B

This can all be programmed using either the GPIB VI or with the VISA VIs in LabVIEW.

Solution for D-Mode:

In MAX:

In the GPIB Interface Properties, Under termination, set the "Terminate Read on EOS"  and then select the proper "EOS Byte"

In LabVIEW:

You will need to use VISA to communicate if you want to programatically change the termination character settings. The settings that need to be modified are,
VI_ATTR_TERMCHAR and VI_ATTR_TERMCHAR_EN setting the termination character and enables status respectively

I am currently placing this process in a Knowledge Base Article, and will post a link with pictures once it is complete.



-Marshall R
0 Kudos
Message 4 of 5
(3,695 Views)

Thank you very much for your help - it finally works 🙂

It looks like the main issue was the EOI.

Again - thanks for your effort.

EE

 

0 Kudos
Message 5 of 5