Hello,
I'm assuming that you're using the instrument driver from National Instruments' site for the TDS 1002. I found it here:
http://zone.ni.com/idnet97.nsf/9b2b33e1993d877786256436006ec498/20152aede3a9e06786256bfe005ff2df?OpenDocument
According to the TDS 1000/2000 programmer's reference that I found on Tektronix's website (www.tek.com), the "CURVe" command will return data that has been captured by the oscilloscope. From what I can tell by a brief reading of this manual, it seems that the CURVe and WAVFrm commands will return nothing if nothing is displayed on the screen.
I believe that you will have to capture the data, display it on the screen, then transfer the data to LabVIEW. It does not seem like it is possible to configure the TDS 1002 to c
ontinually output data points for reading by your computer without prompting, if this is what you are looking for. I suggest that you contact Tektronix to be sure, but based on my reading of the manual, this will not work.
If you are looking to acquire a number of data points and bring them into LabVIEW, you should take a look at our data acquisition products here:
http://www.ni.com/dataacquisition/
Thanks,
Scott B.
Applications Engineer
National Instruments