GPIB SCPI Communication with Agilent E3631A - Invalid separator error
I'm using a NI PXI-8106 Controller including a GPIB-Connector to communicate with a Agilent E3631A power supply. Therefor I downloaded the most recent Agilent Device Driver I obtained through the LabView "Getting Started"-Screen-Option "Tools->Instrumentation->Find Instrument Driver".
For testing I made a simple VI, initialising GPIB communication (GPIB0::05::INSTR), setting output 1 to 1V, enabling output and closing the communication port using the Agilent VI's. This always resulted in error -103 at the power supply, meaning "invalid separator", which first occures when the "*IDN?"-SCPI-Command is sent in the Initialize-VI from Agilent.
No problem at all occured while communicating with the power supply via RS-232, using the same VIs. (Which, apparently, isn't what i want to do ;))
I tried to pinpoint the problem by rebuilding the Agilent-VI using LabViews VISA-VI's. It seems like the "*IDN?"-String (and any other SCPI-Command) is sent in some "faulty" way by the LabView VISA Write-VI when using GPIB.
NI-Spy-Protocol using GPIB:
    Description                                                    Status
 1  viParseRsrc(0x029F0C58, "GPIB0::5::INSTR", 1, 0)               0
 2  VISA Open ("GPIB0::5::INSTR", False, 0, 0, "GPIB0::5::INSTR")  0
 3  VISA Set Attribute ("GPIB0::5::INSTR", TMO_VALUE, 10000)       0
 4  VISA Get Attribute ("GPIB0::5::INSTR", INTF_TYPE, 1)           0
 5  VISA Clear ("GPIB0::5::INSTR")                                 0
 6  VISA Write ("GPIB0::5::INSTR", "*IDN?")                        0
 7  VISA Read ("GPIB0::5::INSTR", 256)                             0xBFFF0015
Agilent GPIB (IEEE-488)-Settings: Adress 5 (it is only possible to change the GPIB Adress, everything else is fix)
PXI GPIB-Settings:
GPIB Interface ID: GPIB0
Primary Adress: 0
Secondary Adress: none
System controller: True
I/O Timeout: 13 (10sec)
Autopolling: True
Send EOI at End of Write: True
Terminate Read on EOS: False
EOS Byte: 0
8-bit EOS Compare: False
SET EOI with EOS on Write: False
I also implemented a wait-cycle of 100ms between each step (VISA Open, VISA clear, VISA write, VISA read) as it was mentioned as a possible solution to a GPIB communications problem elsewhere in this forum, but that also didn't help.
Anyone got an idea how to fix this?
Many thanks in advance,
Michael
(7,600 Views)
How embarassing... problem resolved Smiley Surprised
Someone from another department was kind enough to borrow me his E3631A Power Supply aaand - it's a miracle - it works.
Seems like the last repair at Agilent fixed the defective adjust-knob, but had some effect to the GPIB interface...
Well... easy fix... at last
(7,593 Views)
dear Micheal,
as u told u r using 8106 controller for communicating with agilent power supply.
can you please let me know what type of GPIB cable u used.
is there is any NI part number for that?
usually normal GPIB cables will be bulk head connector. in 8106 it's a small connector.was there is any adaptor attached to the controller?
you can mail me the info, if you have any to my id.
thanks in advance.
VINAY KUMAR D I
(7,389 Views)
When asking a new question, it is recomended that you start a new thread and not post to an older, unrelated one.
All you have to do is do a search for PXI-8106 and you will find the product page at https://www.ni.com/en-us/support/model.pxi-8106.html.
Look under the accessories for cables.
(7,376 Views)
thanks for the info.
i found the accessory.
VINAY KUMAR D I
(7,362 Views)
