GPIB Control Scope VDIV Problem
can anyone help me on the issue I met for GPIB control? The romote controlled instrument is LeCroy waverunner LT584M. I programmed the code ibwrt(scope, "C1:VDIV 2V", 10); to try to set scope volt div as 2v/div. however the scope was set as 20V/Div.
whatever what value I pass to scope to set volt div, the scope always converts the value 10 times larger. While when I try TDIV any other setting commands, so such problem happened.
Anyone know why?
CC
(3,894 Views)
Anyone know why?
CC
What does the programmer's manual say?
If indeed this is a bug with the instrument's firmware, you need to contact LeCroy directly.
(3,876 Views)
It almost sounds like you set the probe to be a 10X.
(3,872 Views)
It sounds EXACTLY like that.  The V/Div setting on those scopes does not adjust for probe gain.
"Should be" isn't "Is" -Jay
(3,864 Views)