03-20-2020 12:36 PM
Hello folks, I am trying to establish communication with a VACOM MVC-3 Vacuum gauge. The manual for the gauge can be found at:
https://www.vacom.de/en/downloads/category/775-total-pressure-measurement?download=2440:mvc-3
I am having problems understanding how to communicate with the device. I am using the built in COMM port (Port 1) of my computer (Windows 10 Enterprise, 64 bit, version 10.0.18363 build 18363) and I have configured the device to have a baud rate of 9600 and to use RS232 as described on page 45 of the manual.
I am trying to communicate with the device in NI Max as I get the feeling from these pages that this is the best way to start..
In Visa I set the Baud rate =9600
data bits = 8
stop bit=1
parity=none
flow control= none
I enabled a termination character.
End mode for reads=TermChar
End mode for writes = Termchar
Termination character= carrage Return \r
Now I started trying to communicate with the simplest command possible , RVN<CR>. (to read the version number P 52 of the manual) and I always get a read error Visa: Hex 0xBFFF0015 Timeout...
In the manual I noticed that they want you to enter the commands in ascii. Does that mean I should convert RVN<CR> to 82 86 78 0x0D or something like that?
Should I write the command and just read, or is there some kind of acknowledgement command that needs to be sent first. (I get vibes from this from page 49 but I am not sure)
Also, since Visa is configured to use \r format as the termination character, does this pose a problem for communicating with this kind of device?
Thank you in advance people.
Happy Weekend
DrRob2020
03-20-2020 01:41 PM
You should be sending "RVN" followed by a Carriage Return. You can configure VISA to send it for you (Send End Enable and Send End Mode I think are the properties you need, I don't use them).
03-21-2020 09:10 AM
Hello crossrulz
Thank you for your quick response. I had attempted this before but it is not working. Am I doing something stupid? Also are there restrictions on how long serial connections can be before signals degrade? I am currently using 10 m (2*5m connected by gender changers) of shielded serial cable:
https://se.rs-online.com/web/p/serial-cable-assemblies/1116151/?relevancy
Thanks again
DrRob2020
03-21-2020 10:20 AM
03-21-2020 01:23 PM
So quickly going through the manual, I see a few things to watch out for.
1. Make sure the internal jumper for the controller is set up for RS-232. There is an option to set it to RS-485.
2. Make sure you are using the same baud rate as the controller. It could be 9600, 19200, or 38400 with the default being 19200.
3. The fact that you used a gender changer scares me. You should not have to do that. If you do, you likely have a NULL Modem cable on one side, which would cause your RX and TX lines to be swapped and you will not talk. Buy a cable that meets your length requirement, is a straight through cable, and has a male on one side and a female on the other side.