Can USBTMC devices be accessed over a network using nivisaserver on Scientific Linux?
I have v15.1 of NI-488.2 & NI-VISA running on Scientific Linux 6.7 on a 32-bit thin client computer. This computer has a PCI-GPIB adaptor installed and has several network/serial/USB ports. This allows all the equipment I have to be accessed over the network using VISA Remote Server.
I've just connected a USBTMC device (a Tektronix AFG3022C), but this doesn't appear under the list of devices being exposed by the NI-VISA server when viewed in MAX from my desktop PC.
Looking on the thin client, I see the device appears under \dev\usbtmc0 and doing an lsusb -v, I see the following:
Bus 002 Device 002: ID 0699:034a Tektronix, Inc.
Device Descriptor:
bLength 18
bDescriptorType 1
bcdUSB 2.00
bDeviceClass 0 (Defined at Interface level)
bDeviceSubClass 0
bDeviceProtocol 0
bMaxPacketSize0 64
idVendor 0x0699 Tektronix, Inc.
idProduct 0x034a
bcdDevice 0.42
iManufacturer 1 Tektronix, Inc.
iProduct 2 Tektronix AFG3022C
iSerial 3 C021006
bNumConfigurations 1
Can nivisaserver (v15.1) be configured to enable the exporting of USBTMC devices over the network, just as it currently does GPIB and serial devices?
Alex
(2,849 Views)
I have answered my own question!
I had to blacklist the native usbtmc kernel module, so that the NI drivers could instead claim the device and share via nivisaserver.
To do this, simply blacklist the module by editing /etc/modprobe.d/blacklist.conf and add "blacklist usbtmc" (as root).
If the driver has already been loaded, you'll also need to "rmmod usbtmc".
The USBTMC device should then show up in NI-MAX.
Alex
(2,801 Views)
