5-volt output usb-8451 vs 8452
I'm trying to clarify the specs for the 5-volt out on the usb-8452.  The 845x hardware/software manual says the module can only supply up to 20 mA.  Yet, the USB-8451 has a maximum of 230 mA.
Is there a mistake in the 8452 manual or did NI cut the available current to 1/10 of the original SPI module?
I ask because we have several 8451 in our testers using the +5v out with a draw of under 100 mA.  Future testers and replacements of the 8451 would mean adding another power supply to the system.
Just thought I'd better check.  Thanks!
(1,730 Views)
Accepted by topic author mysticfree
On pages 3-11 and 3-21 of this  manual, compare the "+5 V Power Source" indications for both modules.
The 20mA limitation of the USB-8452 is unfortunately not a mistake.
(1,694 Views)
I’m currently using the NI USB-8452 for SPI communication, which operates at a 3.3V logic level. However, the peripheral device I’m interfacing with requires 5V logic for proper SPI signaling.
Is there any way to configure or modify the USB-8452 to output 5V I/O levels directly, without using an external level shifter or voltage translator?
Ashok G
(101 Views)
Ashok G
What is your device spec and why a 5V TTL logic device not happy with a 3.1V VOH from USB-8542 SPI?
Unfortunately, you need to work with 5V CMOS, then you need external level shifter, no other option.
Santhosh
Soliton Technologies
(91 Views)
