35660A -Command requires GPIB Controller to be Controller in Charge

Hello
 I am trying  to read and writeHP 35660A siganl Analyzer through a GPIB.

This same VI worked with another signal analyzer , but with this one it shows 2 error msgs.. see atchements.

any ideas?

Nghtcwrlr

---------------------------------------------------------------------------------------------
*************************************
---------------------------------------------------------------------------------------------
Download All
0 Kudos
Message 1 of 4
(3,906 Views)
If the program works with another 35660A, then I would suggest you check to see if the firmware on both models are the same.
0 Kudos
Message 2 of 4
(3,899 Views)

Hey Dennis

 

 I think the GPIB interface of this device is defect. Cos i tried to connect another signal analyzer with the same cable and its been recognized by the Measurement and automation.

 

 

So am planning to change to the other analyzer.

 

Thankss

Nghtcwrlr

---------------------------------------------------------------------------------------------
*************************************
---------------------------------------------------------------------------------------------
0 Kudos
Message 3 of 4
(3,875 Views)

Hi

Certain function resquies that GPIB board to be the CIC (Controller In Charge). That could be your issue. Look at the following kb

- What Is the CIC Protocol in GPIB?

- GPIB Error Codes and Common Solutions (Part 1)

 

Best regard

Nick

 

0 Kudos
Message 4 of 4
(3,864 Views)