11-16-2006 09:58 AM
11-16-2006 10:56 AM
11-16-2006 12:01 PM
11-16-2006 12:18 PM
11-20-2006 01:19 PM
Thanks. I have implemented the Reference Trigger as you described and notice significant improvement. If I acquire many records, say 50, all are grouped very tightly, without the 20 ns jitter previously noted. Now the jitter seems to be about 5 ns or less.
After I fetch all data from the digitizer it is then re-armed via software control. I seem to lose sync at that point. The next group of 50 records are all tightly aligned relative to each other, but slightly off from the previously acquisition. I am confident it is not a function of the pulser / receiver unit.
Now I want to sync the PXI 5124 200 MHz sample clock to the PXI 10 MHz clock (the PXI 6608 card in slot 2 is the source of my trigger pulses). I am not sure exactly how to use the NiScope Configure Clock.vi. For the PXI 5124, do I only need to specify the source to syns to (PXI 10)? What about the 'sync pulse' input parameter? Does it pertain to the PXI 5124?
In other words: If I sync to the PXI 10 MHz signal, do I need to re-sync just after re-arming the 5124? On what line does the 5124 receive the PXI 10 MHz signal? What triggers synchronization? It seems to go out of sync after acquisition is complete, and another acquisition is initiated.
11-21-2006 11:05 AM
11-21-2006 11:10 AM
What about the 'sync pulse' input parameter? Does it pertain to the PXI 5124?
11-21-2006 11:45 AM
The Sync Pulse parameter is used when synchronizing older devices such as the NI 5102 and the NI 5112. It does not apply to the NI 5124. The following is an excerpt from the High Speed Digitizers Help document:
"For the NI 5102, specifies the line on which the sample clock is sent or received. For the NI 5112/5620/5621/5911, specifies the line on which the one-time sync pulse is sent or received. This line should be the same for all devices to be synchronized."
I hope this helps,
Sean Close
Product Support Engineer: High-Speed Digitizers
11-21-2006 01:09 PM
Thank You.
I can clearly see a difference when I sync the sample clock to the PXI_CLK10.
What more can I do beyond synchronizing the 5124 sample clock with the PXI_CLK10? Jitter is still in the 'nano-second' range. How do I get closer to the 'pico second' range spec?
This jitter is present when triggered via a 6608 generated pulse routed to the PXI Trigger line, and also when I use an external trigger (an electrical split of the trigger I am presenting to a pulser/receiver unit). Knowing this, I am assuming it truly is a trigger jitter problem.
Any feedback would be great.
11-21-2006 01:56 PM