How to use 5105 to capture 1second data at 40MHz / Channel , for 4 channels

Dear all,
 
I am using a 5105 SCOPE to try to capture 1second signal at 40MHz , for 4 channels.
So of course, i send the Min. Record Length to 40M & Min. Sample Rate to 40MHz,
But i got a failure result "Out of memory XXXXX". But if i lower the Record Length to say, 100K,
the operation is OK.
 
So, anyone can help me?
0 Kudos
Message 1 of 2
(6,382 Views)
How much memory does your scope have?  Four channels at 40MHz for one second is 4 * 40M * 1 * 2byte/sample = 320MBytes.  Unless you have the highest memory option (512MBytes), you will not be able to store it all in the scope's memory.  If you have the medium memory option, you may be able to get the job done by streaming data from your device as you take it, although even that will be iffy (PCI bandwidth is practically about 100MBytes/sec or less, depending on system load).  Options are to drop your acquisition rate or your time length, neither necessarily good.  You may be able to drop your resolution from 2 bytes to one, but it has been so long since I used the NI-SCOPE driver and I have never used the 5105 that I don't know if this is even possible.  This would only work if you have the 256MByte memory option.