FieldPoint Family
load cell resolution
Can someone help me calculate the smallest change in weight I will be able to read with the cfp-sg-140 and a 10 lbs load cell with a 10 volt excitation and 3mV/V output. I know this module is 16 bit 2^16, but how do I translate this to pounds or grams??? How about if I use a 5 lbs load cell??
Thanks a million in advance for your help....
(9,136 Views)
Have a look at page 9 of the following document:
FP-SG-140 and cFP-SG-140 Operating Instructions
Adnan Zafar
Certified LabVIEW Architect
Coleman Technologies
(9,134 Views)
The manual explains the scaling procedure but I dont have any doubts on that.
I actually meant the smallest change that the cfp-sg-140 will detect with that kind of load cell. Do you know the resolution that I will be able to get out of such configuration????
(9,123 Views)
Just do the math.
 You would need the 3.5 mV range for a 3 mV/V input.  1/2^16 x 3.5 mV/V is 1 part in 65536 is .0534 microvolts/V
Take that times 10lbs divided by 3 mV/V = 1.78 e-4 lbs
(9,118 Views)
