From your gram and mV/V resolutions above it looks to me like your gages are
giving you about 0.433mV/V at 5kg; sounds very small. However, if you calculate
back, assuming the +/- 3.5mV/V input range and that gage sensitivity, your
gram resolution should be about 1.23 g so I think the .433mV/V sensitivity may
be true.
All that being said, I think it will be a real challenge to get 0.1g accuracy from your
gages. We have spent many months working closely with a competent gage
manufacturer to get a 0.1g accurate measurement from 300g gages that give us
~ 2 mV/V @ 300g. Your gages are many times less sensitive than the ones we
are using. Our system yields a gram resolution of ~ 0.02 g.
My suggestion would be to get a gage that gives you as close to one of the standard
SG-140 input ranges (+/- 3.5mV/V, +/- 7 mV/V, etc.) as possible at 150 g. For example,
a 3 mV/V @ 150g gage will give you a gram resolution of 0.005 g. Much easier
to get a specific accuracy if your resolution is smaller
😉Hope this is useful.
Matt