Overview
This example shows the LabVIEW code for programmatically verifying and oscillator frequency accuracy of the NI 5402/5406
Description
This example shows the LabVIEW code for programmatically verifying and oscillator frequency accuracy of the NI 5402/5406 as described on page 11 of the NI 5403/5406 Calibration Procedure.
Requirement
Software:
Labview 2012 (or compatible)
NI-Fgen driver which is compatible with your hardware and version of LabVIEW
Hardware:
NI 5402 or 5406 device
Complete this test to verify the frequency accuracy of the oscillator on the NI 5402/5406. In this test, you generate a 10 MHz sine wave with the NI 5402/5406 and measure the sine wave frequency with a frequency counter.
| Calibration Test Limits ±4.5 ppm | Published Specifications ±25 ppm | ||
| Low | High | Low | High |
| 9,999,955 Hz | 10,000,045 Hz | 9,999,750 Hz | 10,000,250 Hz |
Block Diagram
Description-Separate-2Example code from the Example Code Exchange in the NI Community is licensed with the MIT license.