Example Code
Frequency Counter Measurement Using 5661
Code and Documents
This example shows how to use RFSA to perform frequency counter measurements. The measurements are highly dependent on the clock timebase. Since NI PXI-5661 is equipped with an OCXO, the readings have very low variance.
The example contains two tabs: One that acts like a spectrum analyzer and one that displays frequency measurements. You need to use first the spectrum tab (Figure 1) to find the highest peak you want to measure frequency from. Once the peak has been found, you can change the tab to visualize the frequency with great accuracy.
  1. Set the "Resource Name" (device name under DAQmx Devices)
  2. Make sure the source for the reference clock is set to "Onboard Clock" (or use an external reference if so desired)
  3. Set the "Reference Level (dBm)"
  4. Run the VI.
  5. Set your "Center Frequency"
  6. Set your "Span (Hz)"
  7. Set your "RBW (Hz)" (if you need more/less resolution to visualize your spectrum)
Figure 1. Spectrum Tab
8. Once you see a detected peak on the spectrum tab, switch to the "Frequency Counting" tab.
9. Set the "Integration Time (s)". See Figure 2.
Figure 2. Frequency Counting Tab
The graph shows how the measurement varies in time and gives the jitter (Hz rms) of the measurements.
John Passiak
Contributors