Synchronized data delayed between two PXIe-4464 cards

All,

 

Could you please share any insight or thoughts on a data delay between two PXIe-4462 cards?  Both cards are in one measurement task through channel expansion, sampling at the same rate. One common signal is fed into Channel 0 in both cards. The screenshots below show the delay between the red and yellow traces. 

 

This article give a few options to address the issue. I thought the first option through channel expansion would work. 

 

Thanks,

XR

 

 

 Screenshot 2021-07-19 153125.pngScreenshot 2021-07-19 153249.png

0 Kudos
Message 1 of 5
(3,556 Views)

Please attach your code. Also, what versions of DAQmx and LabVIEW are you running? Do you have two 4462 devices, two 4464 devices, or one of each?

Doug
NI Sound and Vibration
0 Kudos
Message 2 of 5
(3,522 Views)

Hi Doug,

 

I am using LabWindows 2017 with DAQmx version 19.1.1. I have two PXIe-4464 cards. Here is the code that I think is relevant: 

 

/* create the task */

DAQmxCreateTask ("", &probe_TH);

 

/* Setup the AI voltage channels using channel expansion. In total we have eight channels from two cards, four from each card. */

for (i = 0; i < MAX_4464_CHAN_NUM; i ++) {

      DAQmxCreateAIVoltageChan (probe_TH, (ch_info+i)->src, (ch_info+i)->name, DAQmx_Val_Diff, Vin_min, Vin_max, DAQmx_Val_Volts, NULL);
      DAQmxSetAICoupling (probe_TH, (ch_info+i)->src, DAQmx_Val_DC);
      DAQmxSetAIFilterDelayUnits (probe_TH, (ch_info+i)->src, DAQmx_Val_Seconds);
      DAQmxSetAIRemoveFilterDelay (probe_TH, (ch_info+i)->src, (bool32) 1);

}

 

Thanks,

XR

0 Kudos
Message 3 of 5
(3,519 Views)

Sorry for the typo in the first post: there are only two PXIe-4464 cards, no PXIe-4462 card. 

0 Kudos
Message 4 of 5
(3,518 Views)

Here is what I see using the DSA Soft Front Panel:

dsbNI_0-1626928264167.png

 

The 4464 channels are added to the same analog input task. The synchronization is well within the interchannel phase mismatch for a single device. I see similar results when running the LabVIEW example: DAQmx\Analog Input\Voltage - Finite Input.vi

 

Given that the SFP and the LabVIEW example demonstrate synchronization via channel expansion and no explicit synchronization code, I recommend taking out code to remove filter delay.

 

>     DAQmxSetAIFilterDelayUnits (probe_TH, (ch_info+i)->src, DAQmx_Val_Seconds);
>     DAQmxSetAIRemoveFilterDelay (probe_TH, (ch_info+i)->src, (bool32) 1);

 

Doug
NI Sound and Vibration
0 Kudos
Message 5 of 5