Digital Multimeters (DMMs) and Precision DC Sources
Error when trying to adjust PXI-4110 using Calibration Executive
Need help!!
When trying to adjust the current output accuracy on a PXI-4110 using Calibration Executive I received the following error:
Error  -300010  occurred at GetUserFieldData.vi
Possible reason(s):
Calibration Executive: Invalid Data Type in UserFieldDataType. Must be 'NUM' or 'STR'.
Complete call chain:
cex_UtilityGetUserFieldData.vi
cex_UtilityGetUserFieldDataByIndex.vi
_NI 4110_GetChannelRange and Level.vi
Adjust_NI 4110_Current Out Accuracy.vi
at step Adjust Current Output Accuracy
I am using Calibration Executive version 3.2.2, calibrating with an HP 3458A DMM and using Clarostat 240-C Power resistor Decade box.
(8,059 Views)
Hi Rick,
Thanks for posting the error code. I was wondering if you could also post what current output value you entered into the field? Thanks.
Kent
(8,043 Views)
I didn't get to enter anything, as soon as it tried to start the current adjustment portion the error window popped up. It went through the Voltage adjustment portion ok, then I got the error as soon as it switched to the current portion.
Does it actually require user input? I assumed it would be automated like the voltage adjustment part.
(8,040 Views)
Rick has contacted National Instruments directly--this forum post will be updated with a solution when this issue is resolved.
Cheers,
Brooks
(8,006 Views)
Hello ALL,
This error was reported to R&D (CAR# 122736) for further investigation. There are two possible work arounds at the moment.
1. Use the HP 3458A DMM in the unsupported instrument mode. This means that the measurements or DMM readings will have to be entered in manually.
2. Use the NI PXI-4071 for calibration.
Thanks to all for the feedback! Rick thanks for the information and working with us.
Cheers!
Corby_B
http://www.ni.com/support
(7,813 Views)
