Digital I/O
input capacitance of CLK IN for PXI-6551
For the PXI-6551, what is the input capacitance of the CLK IN pin on the front panel of the 6551?  If I were to supply an external clock using an off the shelf oscillator IC this information is important.  I would like to pick a 8.192MHz oscillator to drive CLK IN.
Please advise as to what capacitive load it would normally present to an oscillator.
Thanks.
Anand
(3,706 Views)
Anand,
This is not something we specify for the NI 655x devices.  The CLKIN input to the 655x devices is AC coupled using a 1uF capacitor.  Your signal will hit this DC block before it reaches the configured 50ohm/1kohm input impedance.  If you measure the capacitance at the SMB connector with an LCR meter, you'll be measuring this blocking cap.
(3,692 Views)
Thanks Ryan,
That helps.  So the load seen would be either 50Ohms or 1kOhms.  The oscillator module can drive 50pF at 8.192MHz which is ~ 388 Ohms.  So I should configure CLK IN with 1kOhm input impedance as adequate.
Anand
(3,684 Views)
Hey Anand,
I would recommend seeing which setting works best for you.  Regardless you will see slight amounts of reflection either way so you can just see which results in the best measurement.
David Pratt
Group Manager for Application Engineering Specialists | Automated Test
NIC
(3,647 Views)