04-13-2010 12:19 PM
Hello,
I am using a CDAQ chassis 9172 with a 9423 module and a 9425 module.
Can I make correlated measurements with the 9423 AND static reading with the 9425 at the same time?
I am thinking of creating 2 separate tasks for this. One task for the corellated measurements and one for the static. Would this work?
Thanks
Dan
Solved! Go to Solution.
04-13-2010 04:47 PM - edited 04-13-2010 04:47 PM
Absolutely. With the cDAQ-9172 you can have 1 correlated digital input task, 1 correlated digital output task, and any number of static digital tasks running at the same time.
On the cDAQ-9172 the modules in the correlated digital tasks need to be in slots 1-4. There is no slot restriction on static digital tasks.
Just as heads up, the NI 9425 can currently *only* do static input, but that looks like what you want.
04-13-2010 06:48 PM