Hello,
What is the operating temperature range for the myRIO?
Can it be used in sub-zero temperatures (-20C, 0C)? The plan is to use it to log temperatures in a cold laboratory to calculate heat loss.
I found the following specs in the user guide:
Environmental
To meet these specifications, you must operate the NI myRIO-1900 with the window facing away from the mounting surface and ensure that there is at least 1 in. of clearance in front of the window during use.
Ambient temperature near device (IEC 60068-2-1, IEC 600682-2).......................0 to 40 °C
Storage temperature (IEC 60068-2-1, IEC 600682-2).......................-20 to 70 °C
But what specifications will be affected by colder than 0deg C?
Thank you for the help!
Solved! Go to Solution.
11-01-2015 09:35 PM
Storage temperature is only when the device is powered down.
"These specifications" refers to the specifications document as a whole so if you want to run the myRIO at temperatures below 0C we cannot guarantee it will meet any I/O specifications outlined in the document.