Triggering from multiple trigger sources?

Solved!
Go to solution

Hi,

I have a PXI system with four ADC in it. Also I have four trigger sources one for each card.

In some regimes I need to start all ADC cards simultaneously using that triggter which comes first. I do not know which of the triggers will be first. Is there any ways to realize it in LabView?

Thanks in advance.

0 Kudos
Message 1 of 4
(3,167 Views)

You didn't mention what hardware you are using, but M Series DAQ boards support change detection which could do what you need.

 

If you don't have hardware that supports change detection, you could always use an external OR gate.

 

 

Best Regards,

John Passiak
0 Kudos
Message 2 of 4
(3,153 Views)

thanks for reply.

I'm using S series NI PXI-6123 in four slots of PXI-1044 Chassis.

 

What do you mean under external OR gate? additional hardware?

If the RTSI bus could help me in this case?



0 Kudos
Message 3 of 4
(3,142 Views)
Solution
Accepted by topic author Podoba

Hi Podoba,

 

The 6123 does not support change detection, so the best course of action is probably to use an external OR (or NOR) gate with 4 inputs.  You can connect your signals to the inputs of the OR Gate, then use the resulting signal to trigger your DAQ devices.  If you use a NOR gate, then you would trigger on the falling edge.

 

The "additional hardware" is really is just a simple IC.  You can probably find something that suits your needs on digikey (example search parameters below):

 

        2010-10-22_102002.png

 

 

Best Regards,

John Passiak
0 Kudos