01-30-2008 05:13 PM
01-31-2008 06:19 PM
Hello Moor,
Thank you for using our discussion forums.
There is no way to activate additional input configurations for the PCI-6115. Have a look at the DAQ S Series manual on pages 3-15 and 3-17 for proper connections for non referenced and and ground referenced configurations.
What type of signal are you measuring, and what is its frequency? What exactly do you mean when you say “you see the signal's derivative” when acquiring with the 6115? If you can provide me with a screen shot of what you are seeing it might be more helpful.
The method in which an oscilloscope and a DAQ card acquire and display data is different and if you are just seeing a phase shift of the signal you are reading, you may be acquiring the correct data, however your scope may be triggering at a different point and you might see a phase shift between the data acquired between the Scope and the DAQ card.
Another thought is that if you have a highly capacitive circuit, and there is an impedance mismatch between terminals, then you may be creating an unintended low pass filter, which may be causing the signal you are seeing. So check to make sure that your connections are solid.
Chris_K
02-06-2008 12:28 PM
02-06-2008 05:37 PM
02-06-2008 05:39 PM
02-07-2008 06:25 PM
Hi Moor
Thank you for the screen shots, they were very helpful.
It looks like the signal that you have acquired with the PCI 6115 is AC coupled, and AC couples signal may resemble a differentiated signal. To fix this, while you are setting up you input task with the DAQ Assistant, under Voltage Input Setup, select the Device Tab and choose DC under coupling mode.
This should fix the signal you are acquiring.
Just for reference, here is a link that explains what pseudodifferential is.
Chris_K
02-08-2008 02:56 AM